Ralph Gable
New to nano VNA ... just got the latest and greatest.
I've used port extensions with my higher end Tek VNA and my "miniVNA tiny". I successfully applied port extension with the "VNA View" software with the nano VNA. However, the Saver software ... I calibrate as usual. I set things up to see S11 Phase. Connect my adapter Click on Sweep so I have a reference ... where I am starting Open calibration and enter 500 ps as an offset delay Click on Apply Close the Calibration window. Click on sweep. The phase does not move ==> the Offset Delay does nothing. What am I doing wrong? Ralph WA2PUX YouTube Channel:Electronics for the Inquisitive Experimenter |
On 2022-05-11 15:46:+0000, you wrote:
However, the Saver software ...Maybe nothing :-) There are some things in Saver that don't work well. I havna tried this approach, but have applied e-delay on my -H4, and it works as expected. I take two approaches to using remote software. 1. Cal the device, including e-delays, and tell the software to use that cal data, even if I use multiple sweeps. 2. Reset the device, so that there is no cal, then do the cal in software. It might be that this is the only way the Offset Delay can be applied. Please try that...and let us know. I assume from the lack of responses that no one reading so far, I included, know the answer, and I am not set up to run a test. ~R~ |
Ralph Gable
WOW! Seems like a lot of hoops to jump through to use the Saver software
when measuring impedance. The "VNA View" software does it slick as can be. So, I suppose that the thought is to use the "VNA View" software when measuring impedances or just the stand alone unit ... at least until they get that functionality working properly in Saver. *Ralph* *WA2PUX* *YouTube Channel:* Electronics for the Inquisitive Experimenter On Wed, May 11, 2022 at 3:28 PM Rich NE1EE <TheDustyKey@...> wrote: On 2022-05-11 15:46:+0000, you wrote:However, the Saver software ...Maybe nothing :-) There are some things in Saver that don't work well. |
On Wed, May 11, 2022 at 09:02 AM, Ralph Gable wrote:
I am using NanoVNA Saver 3.10. Offset delay works for me. I calibrated my NanoVNA-H4 device (Not in Saver) from 50 kHz. to 300 MHz. and then attached a 12 inch cable to CH0 (Port 1). You can see the result on the Smith Chart. Then added an offset delay to de-embed for the cable. Smith Chart now shows an open at far right of Smith chart. Roger |
Ralph Gable
I am running the Saver 0.4 version with a nanoVNA V2Plus4 with FW git-20210530-412578c
toggle quoted message
Show quoted text
Ralph WA2PUX YouTube Channel: Electronics for the Inquisitive Experimenter ------ Original Message ------
From: "Rich NE1EE" <TheDustyKey@...> To: [email protected] Sent: 5/11/2022 9:03:16 PM Subject: Re: [nanovna-users] Saver software - Offest Delay seems inoperative On 2022-05-11 18:20:-0700, you wrote:I am using NanoVNA Saver 3.10. Offset delay works for me.Thanks for the update...It might be related to Saver version, because 0.4 is the latest...and it might be related to -H4 firmware version...what are you running? |
Op 12-5-2022 om 04:03 schreef Rich NE1EE:
toggle quoted message
Show quoted text
On 2022-05-11 18:20:-0700, you wrote:I am using NanoVNA Saver 3.10. Offset delay works for me.Thanks for the update...It might be related to Saver version, because 0.4 is the latest...and it might be related to -H4 firmware version...what are you running? |
/Sorry, no text.
toggle quoted message
Show quoted text
/I happened to have the opportunity to measure the same as Roger did when I read this post. Exactly the same measurements made (from 50 kHz to 50 MHz). Calibrated with SDR-Kits BNC Cal Kit. Measurement done with NanoVna -H4 (fw. version 0.5.0) and NanoVna Saver version 0.3.9 Cable used: RG58 A/U Op 12-5-2022 om 04:03 schreef Rich NE1EE: On 2022-05-11 18:20:-0700, you wrote:I am using NanoVNA Saver 3.10. Offset delay works for me.Thanks for the update...It might be related to Saver version, because 0.4 is the latest...and it might be related to -H4 firmware version...what are you running? |
Ralph Gable
I just tried this again ... and it worked.
toggle quoted message
Show quoted text
The difference? I have no idea. As a note ... my high end Tek VNA measured the "test awful load" at 57.91+97.35j (|Z|=113.27). The nanoVNA, with applied delay, reported 57.8+93.2j (|Z|=109.67). This is about a 3% difference. I say this is REALLY good. So ... it looks like we are good to go! Ralph WA2PUX YouTube Channel: Electronics for the Inquisitive Experimenter ------ Original Message ------
From: "PE0CWK via groups.io" <pe0cwk@...> To: [email protected] Sent: 5/12/2022 8:32:19 AM Subject: Re: [nanovna-users] Saver software - Offest Delay seems inoperative /Sorry, no text. |
I ran a quick test this morn. Don't have time to investigate, BUT
nVNA-H4 v4.3_MS, firmware 1.1 Saver 0.3.10 Seemed to work better than 0.4, but still puzzling. On -H4, I corrected by -1750ps, as expected. In Saver, +900ps, even though it displayed the same phase angle to begin with. Saver 0.4.0 Does not connect as reliably as 0.3.10 On both, as I change settings on the H4 and in software, I wind up with index errors, and the only way I can see to get around them is to restart both. Both applied some correction immediately upon changing Offset Delay. The important info is that both software versions report a different delay to offset the same phase. ~20deg CCW phase H4 correction -1750ps both Saver versions, 900ps |
On 2022-05-12 15:32:+0200, you wrote:
Exactly the same measurements made (from 50 kHz to 50 MHz).In your image, I see -1540 ps of correction. The value agrees w a VoP of .66 and .305 m of coax. When do we expect to see a negative angle? I cal'd with ~175 mm extensions, then when I ran the test this morning, I simply didn't add them. I got a +20deg phase, and needed -1750 to correct. But that seems to imply that you did the same, and I kinda imagined that you slapped a 1ft piece of coax on the calibration plane. When I do the same (add some coax), I get a +phase shift, and the H4 needs +1750 of delay. Comment? |
On Thu, May 12, 2022 at 08:46 AM, Rich NE1EE wrote:
I attached 3M of RG-316 to a NanoVNA-H4 and used the TDR function to calculate the round-trip delay. It measured 29.5 nanoseconds. Then the offset delay was adjusted on the NanoVNA until the S11 phase was fairly flat. Offset delay required +29.1 nanoseconds. Screenshot attached. Next the NanoVNA was connected to NanoVNA app. +29.1 ns offset delay was required in the NanoVNA app program. Measurements were done using calibration the NanoVNA app and with NanoVNA calibration. Finally, NanoVNA Saver versions 2.2 and 4.0 were used with calibration done on NanoVNA itself. The offset delay was done in Saver. S11 phase was flat with - 14.56 nanoseconds required which is the one-way delay and negative in sign. Summary - Offset delay was equivalent for all cases but required taking half the round trip delay and negating the sign for NanoVNA Saver. One wonders which one of these methods is used in commercial analyzers...... |
On 2022-05-12 11:57:-0700, you wrote:
H/H4 firmware by default apply internal delay if set and after send it ti CPU. You must reset on device before use Saver software.Thanks...I was using an existing cal, and was resetting the delay to zero before connecting with the software. It makes sense that the delay will be sent to Saver if Saver uses VNA calibration. I was puzzled by 2 things 1. The calibration I /thought/ I did a few days ago is actually off by 20 degrees now 2. Saver did not apply the same offset to get the same sweep to 0 phase degrees I will reset, recal, rerun Saver. Having several ops here report that Saver worked as expected means that I don't have the cal I thought I had. ~R~ |
On 2022-05-12 15:23:-0700, you wrote:
Finally, NanoVNA Saver versions 2.2 and 4.0 were used with calibration done on NanoVNA itself. The offset delay was done in Saver. S11 phase was flat with - 14.56 nanoseconds required which is the one-way delay and negative in sign.thanks for confirming ~R~ |
I did a new cal on the H4. It turned out as before, so the old cal was good.
Here are my notes. H4 phase 1.5 after cal, no cal devices attached. w coax (75 VoP) extension, 1475 ps delay to correct, phase -13.6 After power cycle, phase 1.8 !! connected coax again, phase -17.0, delay 1460 ps to correct So...it seems that I need to cycle power after a cal, and then reload the cal. Hmmm...every time I checked, after the first time, the results were close enough. Saver v0.4.0 - what a mess. Windows 10 x64. phase1.81 Error during sweep Stopped list index out of bounds Manage. Don't know how this actually works. Change to 101 point (H4 is 401), BANDWIDTH to 1000 Hz (H4 is 100) No error this time, so it seems that Saver doesn't like 401 points. S11 phase -17.14, Saver apparently stops H4 from displaying the sweep, so tapping on screen updates. No. Not always. Delay -731 ps - no delay applied in H4 Saver left H4 in unstable state, so cycled power Saver after App H4 phase -17.1, Saver -17.09 Saver delay -730 ps. as before. ** Note here that Saver is still reporting a delay way off from the H4 and App. And that these values are in the ballpark of what I observer previously, so Saver is consistent. I don't know what I am doing different from you folks, but I sure don't get useful delays. Sure the phase goes to zero, but the delay is way off from the H4 and App. App v1.1.208 1.783 After power cycle, no delay on H4, phase 1.789 w coax (75 VoP) extension, H4 phase -17.1 App phase -16.84, delay 1463 ps Disconnected. Closed App. H4 still sweeping. Checked points, delay, bandwidth...all unchanged, BUT Saver changes those in ways I don't understand yet. ** At one point, App went into a loop on access violation. Needed to kill it. Wouldna shutdown. Reran Saver. See above. So at least App is in the same range as the H4. But they should all agree much closer than this. It's all just number crunching. Same numbers, should be same results. If there is some trick to running Saver, I'm open to suggestions. |