2port measurement determining Zdut with S11 and S21


 

Hello Roger,

I thought that the method I used now for the resistors was working. And it
seems to work at least for these 4 resistors (330, 985, 5160 and 21800);-)
But I now did a 1:1 Guanella choke as dut and it seems to behave
differently in a 2 port measurement...
Here are the 2 port measurements NanoVNA Saver of my 1:1 Guanella choke (I
also attached the 11Guanella-corebraid-1s21.s2p file):
[image: afbeelding.png]
As you can see the S21 R+jX has totally different form (e.g. peaks) than
S11 R+jX.

If I use the formula of (Page
10) :
[image: afbeelding.png]
I can determine with the 2 port measurements Zdut (Zx) in two ways through
S11 and S21 2 port measurement.
Z2pseries(S11)=(Zsource+Zload)*S11/(1-S11)
Z2pseries(S21)=(Zource+Zload)*(1-S21)/S21

[image: afbeelding.png]The green curve should be close to the red (blue
dotted or orange dotted) ones (as when using my method for the resistors),
but with the choke they are totally different. I must be doing something
(obviously?) wrong! But what? Any help is welcome.

All the best,

Victor


 

Victor,

I am not sure how you are measuring the DUT using S11 reflection and S21 series methods.

You have to measure them with the correct reference plane calibration and this usually requires two different test jigs. What are you using for S11 and S21 test jigs? Attached is one I use for S11 up to 30 MHz. I tack solder the DUT as shown. My example uses an inductor but you can solder the leads to your choke. The second one I built into a box and it works well too. I posted my S21 series jig in the last topic

Roger


 

Hello Roger,

Using Series though, the NanoVNA measures both a S11 and S21 (for that
configuration, which I [of course] calibrated). The link I gave (
), calculates the Zdut using
these Sx1 from the 2port series through measurement. In the case of the
resistive DUTs the S11 measured with 2 port (red blue dotted) is almost
identical with the S11 measured with 1 port (orange dotted). The same I see
with the toroid DUT.
But I am not able to map the Zdut calculated with the S21 (green dashed;
measured in the series through 2 port).

So the Zdut (Zx) using with 2port series through and S11 based is different
from Zdut (Zx) using (the same) 2 port series through and S21 based. So all
stray capacitances, etc are the same (which I see part of the DUT).

All the best,

Victor

Op za 11 jun. 2022 om 07:26 schreef Roger Need via groups.io <sailtamarack=
[email protected]>:

Victor,

I am not sure how you are measuring the DUT using S11 reflection and S21
series methods.

You have to measure them with the correct reference plane calibration and
this usually requires two different test jigs. What are you using for S11
and S21 test jigs? Attached is one I use for S11 up to 30 MHz. I tack
solder the DUT as shown. My example uses an inductor but you can solder
the leads to your choke. The second one I built into a box and it works
well too. I posted my S21 series jig in the last topic

Roger






 

Op za 11 jun. 2022 om 08:32 schreef Victor Reijs via groups.io
<pe1atn.victor.reijs@...>:

Hello Roger,

Using Series though, the NanoVNA measures both a S11 and S21 (for that
configuration, which I [of course] calibrated). The link I gave (
), calculates the Zdut using
these Sx1 from the 2port series through measurement.
By the way, NanoVNA Saver uses also these formulas to plot the S11 and S21
for the 2port series through measurements (Sx1 R+jX [series]). IMHO these
two should be quite the same, but they are in my case quite different (as
you could see in the earlier screen grab of NanoVNA Saver.

All the best,
Victor


 

Victor,

Let me preface my comments by saying that I have built several test jigs for measuring impedance using the S21 series method and tried several ways to compensate for errors. None have been very successful and I now just use the S11 reflection method which works up to several thousand ohms of complex impedance with a decent test jig.

I believe the NanoVNA-H4 is just not the right tool for the job for several reasons that I have mentioned before....
- the source and termination impedances are not 50 + j0 ohms and vary with frequency
- the current firmware and PC programs only use 6 term error correction not 12 term

Others have found similar issues using the S21 series method using the NanoVNA and even with higher performance HP VNA equipment. Here are some interesting links. The K6JCA experiments find differences similar to what you have measured.




In my experiments I found that parasitic capacitance had considerable affect on the S21 measurements and I tried several different test jigs to see what happened. The issue of where to establish the reference plane took some consideration. S21 Gain and S21 phase determine the complex impedance of the DUT and even with some type of source (Port1) and termination (Port2) correction the phase was often incorrect. Phase measurement was the problem and in my tests was due to an offset delay across the jig. Another issue I found was that for some reason high DUT impedance resulted in common mode current on the surface of the coax cables connecting the NanoVNA to the test jig. I tried snap-on ferrites and the results changed.

You mentioned that you could do a S11 reflection and S21 series measurement at the same time. You will get different results when the DUT is in series between Port 1 and Port2 compared to when it is across Port 1 alone. Best case is that it will be 50 ohms higher. TIn reality the complex impedance measured when in series is the DUT plus the input impedance of the cable attached to Port 2 (which will deviate from 50 ohms due to Port 2 impedance not being 50 ohms) and any parasitic inductance or capacitance between the DUT and the Port 2 cable.

I wish you good luck in your quest and hope you find a method that works for you and that you find a solution that eluded me.

Roger


 

On Sat, Jun 11, 2022 at 11:08 AM, Roger Need wrote:

In my experiments I found that parasitic capacitance had considerable affect
on the S21 measurements and I tried several different test jigs to see what
happened.
Here is one of theS11/S21 test jigs I tried. It could use improvement but I don't have the time right now to continue with this project.

Roger


 

Thanks Roger and Owen,

Remember that the S11 formula from
and NanoVNA Saver (and my own
Excel spreadsheets) already takes away the 50ohm (Zload) of port 2.
The S11 2-port series through measurement (also described by k6jca,
NaonoVNA Saver and earlier link), is always close to the S11 1-port
measurement.

My calibration plane is at the end of the middle/earth tips of the PCB SMA
connector (also my calibration kit is at these tips:
). I solder the
DUT directly to these pins aka calibration plane (so no real extension of
the plane). This all to minimize variability of reference plane
capacitance/etc.

In my case Zload of port2 is very close to on average 50.8-0.1i ohm (300kHz
to 30MHz), which is based on a calibration load of 49.7 ohm (using DVM,
accuracy 1.5%). As my Calibration-load is a little smaller than 50~0iohm,
the resistances measured by the NanoVNA will be a little too high. So my
Zload comes closer to 50ohm. Anyway I compensate S21 for that Zload.
In my case Zsource is significantly different (on average 43.6+3.2iohm)
from 50ohm. So I compensate S21 for that (using Owen's method:
), S11 is not affected by this Zsource
as Zsource is calibrated 'out' by Calibration-load.

I put the DUT on a styrofoam underground (to minimize the effects of my
wooden table top's dielectricum). I also put a clip-on ferrite on the port1
or port2 wire, or shortened the port2 cable (from 16cm to 2cm), or put it
as far away from other equipment (the NanoVNA is though always some 16cm
nearby: IMG_1643.jpg). All these changes don't really have an effect on
2port series through measurements of S11 or S21. Only slightly at the end
of my measurement range (30MHz).
The DUT (in 2 port series through) has *certainly* hand effect (aka adding
stray capacitance), so I need to stay way;-). This hand effect is most
prominent on the S21 (not much on the S11) 2-port series through.

If I change the offset delay, this has almost no effect on S21 2port series
through (as expected), but on the S11 2port series through (it shifts the R
and X peaks to higher frequencies). If I put the offset delay to around
-56ps equivalent to 7.5mm, the R&X peaks of S11 map the R&X peaks of S21.
The below screengrab is with a offset delay of -56ps:
[image: afbeelding.png]
I still find it strange that while the S21 changes so much the S11 stays
quite constant. If I understand k6jca (


), he (as some other experts) likes S21 the best (also more stable than
S11...) for high impedances (like for chokes at say 10kohm). This makes it
even more confusing as others now tell me S11 is better;-)
It would be nice to use Benson's Y method; but that needs a S11, S21, S12,
S22 setup (and with that a 12-term calibration). This is not available with
NanoVNA, so that does not help as advice. Perhaps in a few years prices
will go down for such equipment;-)

I will digest k6jca pages in more detail (I recognise some of what he is
telling, looking at my own experiences).

Thanks for all the feedback provided. If I gain more experience, I will
communicate over the group.

All the best,

Victor

Op za 11 jun. 2022 om 20:08 schreef Roger Need via groups.io <sailtamarack=
[email protected]>:

Others have found similar issues using the S21 series method using the
NanoVNA and even with higher performance HP VNA equipment. Here are some
interesting links. The K6JCA experiments find differences similar to what
you have measured.






In my experiments I found that parasitic capacitance had considerable
affect on the S21 measurements and I tried several different test jigs to
see what happened. The issue of where to establish the reference plane
took some consideration. S21 Gain and S21 phase determine the complex
impedance of the DUT and even with some type of source (Port1) and
termination (Port2) correction the phase was often incorrect. Phase
measurement was the problem and in my tests was due to an offset delay
across the jig. Another issue I found was that for some reason high DUT
impedance resulted in common mode current on the surface of the coax cables
connecting the NanoVNA to the test jig. I tried snap-on ferrites and the
results changed.

You mentioned that you could do a S11 reflection and S21 series
measurement at the same time. You will get different results when the DUT
is in series between Port 1 and Port2 compared to when it is across Port 1
alone. Best case is that it will be 50 ohms higher. TIn reality the
complex impedance measured when in series is the DUT plus the input
impedance of the cable attached to Port 2 (which will deviate from 50 ohms
due to Port 2 impedance not being 50 ohms) and any parasitic inductance or
capacitance between the DUT and the Port 2 cable.

I wish you good luck in your quest and hope you find a method that works
for you and that you find a solution that eluded me.


 

I forgot to add that the NanoVNA measured Impedance (based on the S21 (or
S11) 2port series through):
[image: afbeelding.png]
matches quite nicely the theoretical simulation of a choke:

[image: afbeelding.png]

<the theoretical simulation got one parameter (stray capacitance of the
choke, not of the testing gear!) which was derived by matching the peak in
the simulation with the peak in the measured S21 curve>

I think I am more or less understanding what/how is measured with the
NanoVNA. Will try to make a web page out of that.

All the best,

Victor



Op za 11 jun. 2022 om 08:32 schreef Victor Reijs via groups.io
<pe1atn.victor.reijs@...>:

Hello Roger,

Using Series though, the NanoVNA measures both a S11 and S21 (for that
configuration, which I [of course] calibrated). The link I gave (
), calculates the Zdut using
these Sx1 from the 2port series through measurement. In the case of the
resistive DUTs the S11 measured with 2 port (red blue dotted) is almost
identical with the S11 measured with 1 port (orange dotted). The same I see
with the toroid DUT.
But I am not able to map the Zdut calculated with the S21 (green dashed;
measured in the series through 2 port).

So the Zdut (Zx) using with 2port series through and S11 based is different
from Zdut (Zx) using (the same) 2 port series through and S21 based. So all
stray capacitances, etc are the same (which I see part of the DUT).

All the best,

Victor

Op za 11 jun. 2022 om 07:26 schreef Roger Need via groups.io
<sailtamarack=
[email protected]>:

Victor,

I am not sure how you are measuring the DUT using S11 reflection and S21
series methods.

You have to measure them with the correct reference plane calibration and
this usually requires two different test jigs. What are you using for
S11
and S21 test jigs? Attached is one I use for S11 up to 30 MHz. I tack
solder the DUT as shown. My example uses an inductor but you can solder
the leads to your choke. The second one I built into a box and it works
well too. I posted my S21 series jig in the last topic

Roger










 

Victor, I notice that in these examples you are using the nanoVNA with the
saver software, connected to a computer via USB cable.

When I do a similar setup, I note that the USB cable to the computer can
significantly affect the measurement at HF frequencies. I found that I
had to use a high quality cable and put ferrite beads on it to decouple
that RF path. You may wish to experiment with that to see if it gives you
better stability in your results.

On Sat, Jun 11, 2022, 4:22 PM Victor Reijs <pe1atn.victor.reijs@...>
wrote:

I forgot to add that the NanoVNA measured Impedance (based on the S21 (or
S11) 2port series through):
[image: afbeelding.png]
matches quite nicely the theoretical simulation of a choke:

[image: afbeelding.png]

<the theoretical simulation got one parameter (stray capacitance of the
choke, not of the testing gear!) which was derived by matching the peak in
the simulation with the peak in the measured S21 curve>

I think I am more or less understanding what/how is measured with the
NanoVNA. Will try to make a web page out of that.

All the best,

Victor



Op za 11 jun. 2022 om 08:32 schreef Victor Reijs via groups.io
<pe1atn.victor.reijs@...>:

Hello Roger,

Using Series though, the NanoVNA measures both a S11 and S21 (for that
configuration, which I [of course] calibrated). The link I gave (
), calculates the Zdut
using
these Sx1 from the 2port series through measurement. In the case of the
resistive DUTs the S11 measured with 2 port (red blue dotted) is almost
identical with the S11 measured with 1 port (orange dotted). The same I
see
with the toroid DUT.
But I am not able to map the Zdut calculated with the S21 (green dashed;
measured in the series through 2 port).

So the Zdut (Zx) using with 2port series through and S11 based is
different
from Zdut (Zx) using (the same) 2 port series through and S21 based. So
all
stray capacitances, etc are the same (which I see part of the DUT).

All the best,

Victor

Op za 11 jun. 2022 om 07:26 schreef Roger Need via groups.io
<sailtamarack=
[email protected]>:

Victor,

I am not sure how you are measuring the DUT using S11 reflection and
S21
series methods.

You have to measure them with the correct reference plane calibration
and
this usually requires two different test jigs. What are you using for
S11
and S21 test jigs? Attached is one I use for S11 up to 30 MHz. I tack
solder the DUT as shown. My example uses an inductor but you can
solder
the leads to your choke. The second one I built into a box and it
works
well too. I posted my S21 series jig in the last topic

Roger














 

Thanks Stan for this feedback. Anyfeedback is very welcome.
I put click-on ferrites (25mm long) on port1, port2 and USB cables, they
have (in my setup) no significant effect on the measured S21 (or S11)
2port series through measurements.

Anyway, all feedback is very welcome. Every question/query/remark can help
to pinpoint things more precisely. Thanks.

All the best,

Victor

Op zo 12 jun. 2022 om 05:31 schreef Stan Dye <standye@...>:

Victor, I notice that in these examples you are using the nanoVNA with the
saver software, connected to a computer via USB cable.

When I do a similar setup, I note that the USB cable to the computer can
significantly affect the measurement at HF frequencies. I found that I
had to use a high quality cable and put ferrite beads on it to decouple
that RF path. You may wish to experiment with that to see if it gives you
better stability in your results.

On Sat, Jun 11, 2022, 4:22 PM Victor Reijs <pe1atn.victor.reijs@...>
wrote:

I forgot to add that the NanoVNA measured Impedance (based on the S21 (or
S11) 2port series through):
[image: afbeelding.png]
matches quite nicely the theoretical simulation of a choke:

[image: afbeelding.png]

<the theoretical simulation got one parameter (stray capacitance of the
choke, not of the testing gear!) which was derived by matching the peak
in
the simulation with the peak in the measured S21 curve>

I think I am more or less understanding what/how is measured with the
NanoVNA. Will try to make a web page out of that.

All the best,

Victor



Op za 11 jun. 2022 om 08:32 schreef Victor Reijs via groups.io
<pe1atn.victor.reijs@...>:

Hello Roger,

Using Series though, the NanoVNA measures both a S11 and S21 (for that
configuration, which I [of course] calibrated). The link I gave (
), calculates the Zdut
using
these Sx1 from the 2port series through measurement. In the case of the
resistive DUTs the S11 measured with 2 port (red blue dotted) is almost
identical with the S11 measured with 1 port (orange dotted). The same I
see
with the toroid DUT.
But I am not able to map the Zdut calculated with the S21 (green
dashed;
measured in the series through 2 port).

So the Zdut (Zx) using with 2port series through and S11 based is
different
from Zdut (Zx) using (the same) 2 port series through and S21 based. So
all
stray capacitances, etc are the same (which I see part of the DUT).

All the best,

Victor

Op za 11 jun. 2022 om 07:26 schreef Roger Need via groups.io
<sailtamarack=
[email protected]>:

Victor,

I am not sure how you are measuring the DUT using S11 reflection and
S21
series methods.

You have to measure them with the correct reference plane calibration
and
this usually requires two different test jigs. What are you using
for
S11
and S21 test jigs? Attached is one I use for S11 up to 30 MHz. I
tack
solder the DUT as shown. My example uses an inductor but you can
solder
the leads to your choke. The second one I built into a box and it
works
well too. I posted my S21 series jig in the last topic

Roger


















 

Hello all of you,

Op za 11 jun. 2022 om 22:11 schreef Victor Reijs <
pe1atn.victor.reijs@...>:

Thanks for all the feedback provided. If I gain more experience, I will
communicate over the group.
Here is my web page on my experiences. I think S21 and Y21 methods look
promising using my NanoVNA-H:


All the best,

Victor


 

Hi Roger

Apologies for resurrecting an old post
I was just wondering if you could provide any details on the S11 / S21 test jig, it looks pretty much ideal for the matching tests I would like to try.
I am guessing that this is something you modified yourself & it's not available off the shelf ?
What are the origins of the base board ?

Thx
Brian


 

The jig is made using one section of the commercially available NanoVNA test board. Available on eBay, Amazon and from Chinese sellers. You can find discussions about this board on several posts in this group.