Bad Coax...feedback requested


 

I've calibrated my NanoVNA & connected it to an approximate 350' length of coax with a known good dummy load on the opposite end. The S11 plot is attached...not pleasing at all.
Can anyone on the list say what the cause may be?
I have ideas, but I don't want to sway anyone's opinion..


 

Did you put the load on the output of the VNA? This looks somewhat normal
but the return loss is a bit marginal. I am curious about the return loss
of the load by it self. Do you have the VNA saver program to look at the
line in the TDR format?

On Sun, Jan 5, 2020, 7:28 PM Robin Midgett <K4IDC@...> wrote:

I've calibrated my NanoVNA & connected it to an approximate 350' length of
coax with a known good dummy load on the opposite end. The S11 plot is
attached...not pleasing at all.
Can anyone on the list say what the cause may be?
I have ideas, but I don't want to sway anyone's opinion..




 

Most rf lines like low density foam has a spec return loss of about 20 db.

On Sun, Jan 5, 2020, 7:28 PM Robin Midgett <K4IDC@...> wrote:

I've calibrated my NanoVNA & connected it to an approximate 350' length of
coax with a known good dummy load on the opposite end. The S11 plot is
attached...not pleasing at all.
Can anyone on the list say what the cause may be?
I have ideas, but I don't want to sway anyone's opinion..




 

Additional info...it's a piece of 1.25" hardline, and the response is much worse going up in frequency; ~-5 return loss at 146 MHz.


 

Port 00 or Port 01, that used for one-port measurement of s11, may not be a
good 50+j0 source over your frequency range. You exhibit everywhere better
than 15 dB return loss which is pretty good, but the periodicity of the
measurement is a bit disturbing.

Try recalibrating and remeasuring with a 6 dB attenuator on the s11 port.
That will assure a return loss in excess of 12 dB and stabilize the
impedance of that port.

Dave - W?LEV

On Mon, Jan 6, 2020 at 12:28 AM Robin Midgett <K4IDC@...> wrote:

I've calibrated my NanoVNA & connected it to an approximate 350' length of
coax with a known good dummy load on the opposite end. The S11 plot is
attached...not pleasing at all.
Can anyone on the list say what the cause may be?
I have ideas, but I don't want to sway anyone's opinion..



--

*Dave - W?LEV*
*Just Let Darwin Work*
*Just Think*


 

Hi Clyde,
Thanks for the quick response.
Here's a plot of the dummy load across the same span....looks good to me, but I'm here to learn from others.
I haven't tried VNA Saver yet...would have to spend some time on that; hopefully this week.


 

Thanks Dave, I'll try that soon..hopefully tomorrow evening.

On Sun, Jan 5, 2020 at 06:47 PM, David Eckhardt wrote:


Port 00 or Port 01, that used for one-port measurement of s11, may not be a
good 50+j0 source over your frequency range. You exhibit everywhere better
than 15 dB return loss which is pretty good, but the periodicity of the
measurement is a bit disturbing.

Try recalibrating and remeasuring with a 6 dB attenuator on the s11 port.
That will assure a return loss in excess of 12 dB and stabilize the
impedance of that port.

Dave - W?LEV

On Mon, Jan 6, 2020 at 12:28 AM Robin Midgett <K4IDC@...> wrote:

I've calibrated my NanoVNA & connected it to an approximate 350' length of
coax with a known good dummy load on the opposite end. The S11 plot is
attached...not pleasing at all.
Can anyone on the list say what the cause may be?
I have ideas, but I don't want to sway anyone's opinion..



--

*Dave - W?LEV*
*Just Let Darwin Work*
*Just Think*


 

From: Robin Midgett

I've calibrated my NanoVNA & connected it to an approximate 350' length of coax with a known good dummy load on the opposite end. The S11 plot is attached...not pleasing at all.
Can anyone on the list say what the cause may be?
I have ideas, but I don't want to sway anyone's opinion..

Attachments:
350_ coax with dummy load.png: /g/nanovna-users/attachment/9185/0
==============================

Robin,

You might try a TDR plot and see whether there is an obvious discontinuity?

Cheers,
David
--
SatSignal Software - Quality software for you
Web:
Email: david-taylor@...
Twitter: @gm8arv


 

On Mon, 6 Jan 2020 at 00:28, Robin Midgett <K4IDC@...> wrote:

I've calibrated my NanoVNA & connected it to an approximate 350' length of
coax with a known good dummy load on the opposite end.
The S11 plot is attached...not pleasing at all.
Can anyone on the list say what the cause may be?

VNA measurements require phase stables cables. The ones I use for microwave
use (upto 26.5 GHz) are 24¡± long and cost $5000 a pair.

The changes in phase you will observe with very long cables is going to be
much higher than short cables of the same quality. Flexing, temperature
changes are really going to screw you up with long cables.

When I do critical measurements I try to eliminate cables completely.
Measurements of my VNA calibration standards sold in kits are made

There are some ways to test cables for their suitability for VNA use. I
will add some details later on my website under the FAQ section.



There are other issues too, but they might not be relevant on the NanoVNA.
As the frequency of the source is swept, the detector looks for that
frequency. If there¡¯s a very long cable, by the time the reflection is seen
back at the VNA, the source has changed frequency and its detector is then
looking for higher frequency than what it receives.

The solution on some VNAs is to slow the sweep speed down. Since the
NanoVNA does sweep that fast, it is unlikely to be the problem. Reducing
the span to the absolute minimum you can get away with might help, as the
frequency shift between each point will be much lower.

What¡¯s the temperature stability of your load like? If you take a load from
room temperature and change its temperature much, it is likely to change in
value.


I have ideas, but I don't want to sway anyone's opinion..

I wonder if any of your ideas are the same as mine!

Dave, G8WRB.


--
Dr. David Kirkby,
Kirkby Microwave Ltd,
drkirkby@...

Telephone 01621-680100./ +44 1621 680100

Registered in England & Wales, company number 08914892.
Registered office:
Stokes Hall Lodge, Burnham Rd, Althorne, Chelmsford, Essex, CM3 6DT, United
Kingdom


 

Robert,

It would be desirable if you could also measure the transmission through the
cable, including both magnitude and phase versus frequency. Plus any info
you have about the supposed characteristic impedance and velocity factor of
the cable. Extend the lower frequency limit down to 50 kHz as well.

A second set of equivalent plots but with the nano's scan set for a span of about
10 MHz and a center frequency well up in the VHF regime (~150 MHz would be
fine) might also be pretty informative. These should also include the scan of the
termination that you're using by itself.

The overall shape of the plot is somewhat consistent with two point reflections
of roughly equal magnitude, separated in round-trip time by a bit under 500 nsec.
However, the deteriorating overall return loss at increasing frequency is not really
consistent with that simple model. But if your connections to the hard line were
a bit crude, creating added impedance bumps at the two ends, the curve I've seen
so far would be better-explained.

Could you also send a photo showing exactly how you are physically interfacing
to the hard line from the small cable environment?

What I've seen so far suggests a cable whose impedance is not really 50 ohms,
combined with some other issue such as (perhaps) impedance bumps at the two
ends. The period of ~2 MHz seen in the S11 plot is not consistent with a physical
length of 350 ft- in fact it's off by almost a factor of 2 if one assumes a VF of 0.67.

Dana K8YUM


 

I missed what was actually being tested....
Also, if one is testing cables themselves, is there a finite distant limit?

BR

Tim K4SHF

On Mon, Jan 6, 2020 at 6:28 AM Dana Whitlow <k8yumdoober@...> wrote:

Robert,

It would be desirable if you could also measure the transmission through
the
cable, including both magnitude and phase versus frequency. Plus any info
you have about the supposed characteristic impedance and velocity factor of
the cable. Extend the lower frequency limit down to 50 kHz as well.

A second set of equivalent plots but with the nano's scan set for a span
of about
10 MHz and a center frequency well up in the VHF regime (~150 MHz would be
fine) might also be pretty informative. These should also include the
scan of the
termination that you're using by itself.

The overall shape of the plot is somewhat consistent with two point
reflections
of roughly equal magnitude, separated in round-trip time by a bit under
500 nsec.
However, the deteriorating overall return loss at increasing frequency is
not really
consistent with that simple model. But if your connections to the hard
line were
a bit crude, creating added impedance bumps at the two ends, the curve
I've seen
so far would be better-explained.

Could you also send a photo showing exactly how you are physically
interfacing
to the hard line from the small cable environment?

What I've seen so far suggests a cable whose impedance is not really 50
ohms,
combined with some other issue such as (perhaps) impedance bumps at the two
ends. The period of ~2 MHz seen in the S11 plot is not consistent with
a physical
length of 350 ft- in fact it's off by almost a factor of 2 if one assumes
a VF of 0.67.

Dana K8YUM




 

You might try rapping the connectors on the ends of the line and check for any changes in S11 or S21 response. S21 loss should be less then -0.5 dB @ 30 MHz for 350' of 1.25" hardline.

If connectors are good and cable has not taken a lightning strike in its past life you should be more than 20 dB return loss on S11.


 

What does the VSWR trace look like?

*Clyde K. Spencer*

On Sun, Jan 5, 2020 at 7:28 PM Robin Midgett <K4IDC@...> wrote:

I've calibrated my NanoVNA & connected it to an approximate 350' length of
coax with a known good dummy load on the opposite end. The S11 plot is
attached...not pleasing at all.
Can anyone on the list say what the cause may be?
I have ideas, but I don't want to sway anyone's opinion..